hi

How can I Calculating peak Resolution in OpenLab ChemStation C.01.05 &C.01.06?

Regards,

Dheya

Parents
  • Hello,

    Peak resolution should be calculated for all peaks except the first one in the chromatogram as part of processing the results. If you are using classic reporting, you will need to find a report that outputs that value. The extended performance report has most of the suitability values as part of the report. If you are using intelligent reports, you can add the resolution value you want to your current tables.  

    Marty Adams

Reply
  • Hello,

    Peak resolution should be calculated for all peaks except the first one in the chromatogram as part of processing the results. If you are using classic reporting, you will need to find a report that outputs that value. The extended performance report has most of the suitability values as part of the report. If you are using intelligent reports, you can add the resolution value you want to your current tables.  

    Marty Adams