What, if any, is the best method to search for orphan data in OpenLab CDS?

I'm currently writing an SOP regarding OpenLab administration. Specifically, I'm having trouble determining the best method of identifying orphan data as it relates to review of chromatography data in the GMP laboratory. Orphan data is unusual in that it could indicate that an analyst is intentionally placing data in another location not obvious to the reviewer (e.g. test injections for problematic method run prior to full sequence). There are two ways that I know of to perform a search.

1. Via ECM classic client and performing a filtered audit trail search using user name and/or dates. The populated information doesn't seem particularly useful to a reviewer/auditor IMHO.

2. Via Control Panel/Data Analysis and Search Results. Include results from all users and Search across projects boxes are both checked. This seems like the best route but it seems that it isn't possible to customize the search by date. Although you can use the quick search option. In my case, if I use the quick search option, no data is found.

Any help or insight that you can provide would be appreciated!

Parents
  • Hello,

    I cannot tell what ECM or CDS version you are using and that can make a difference in the search capabilities. Take a look at the attached guide on page 50 for information on searching with ECM 3.x and CDS 2.x. You are not extracting any CDS metadata keys for your data that is why the name, acq operator, acq date and instrument name columns are blank for your search results. It also limits the search metadata in later versions. You should be able to use the quick search by dates with the current user to see what was created in defined time frames. If you need more advance searching of the metadata use the advanced search in the ECM client. Please note to search anything outside the 15 general file metadata fields you will need to setup filter keys on the LCDF structure in ECM. You should at least have the OpenLab CDS 2.x injection data filter keys applied to your results folders. I would avoid adding peak filter keys as that will cause the db to grow unless you need to be able to search against peak results. Also adding keys will not filter any existing data already in ECM, if you need to search data already present in ECM you will need to rekey any structures where you add filter keys. 

    Marty Adams

  • Hi Marty,

    We are using ECM 3.6 and CDS 2.7. Thank you, this information is extremely helpful! I will dig into these resources/examples.

    Best,

    Dan

  • Anyone have thoughts on my last question related to this post? Would this be a situation where I would need to work with an Admin to set up the filters/keys properly in the ECM so that data searches in OpenLab Data Analysis list the proper info related to available metadata?

    Thanks,

    Dan

  • Dan,

    Did you check to make sure the rekey extracted the meta-data? You can check the file properties in the LCDF where you added the filter and extracted the metadata. I assume you were unable to search, or the columns were still empty for the search returns. You can see below an example where I only rekeyed one file in the project with the meta data filter as shown. You can see the search worked for multiple files but only the one file had the information in the columns available. If you are still not getting the results you want open a support ticket with informatics and they can look at the issue. 

    Marty Adams

  • Marty,

    I figured out where I went astray. The desired metadata is now populated when performing a search in OpenLab CDS. Thanks so much for your assistance!

    Best,

    Dan

  • Marty,

    I had one additional somewhat unrelated question regarding ECM. We organize our data in locations by quarter (e.g. Q4 2023). If it is desired to "lock" prior quarter folders, is archival the best way to do this? The intent is to prevent adding new data to these folders.

  • Hello,

    You could archive the data just be aware there is no way to de archive in ECM 3.x, so once archived they will be read only. Also, you can only archive files not the folders themselves. You could remove users access to the folders in ECM but that would require setting LCDF level access and not using global access in ECM. The other option would be to have the users lock the data in CDS when complete. This could be setup as part of an electronic signature workflow if you need more control over the locking step. 

    Marty 

  • Thanks for all of your help so far Marty. It is much appreciated! I noticed that when accessing ECM, users without admin access are able to delete folders. I am not an administrator, but I was wondering how to remove that privilege in ECM. Obviously, it is not a good thing for chemist level users to be able to delete results. Am I correct that this needs to be set up separately in ECM and cannot be done in the OLCP?

    Dan

Reply
  • Thanks for all of your help so far Marty. It is much appreciated! I noticed that when accessing ECM, users without admin access are able to delete folders. I am not an administrator, but I was wondering how to remove that privilege in ECM. Obviously, it is not a good thing for chemist level users to be able to delete results. Am I correct that this needs to be set up separately in ECM and cannot be done in the OLCP?

    Dan

Children
Was this helpful?