Curve coefficient didn't correspond to the average.

Hi, 

First, I want to show how I perfomed my 1 level calibration curve. 

We saw that I asked for the average when several calibration standards are used and contained the same target compound. 

We do have the TFT in each of my 4 calibration standards. 

We can see it at my section filtered for Cal std only. 

What's weird here it's the first RF calculated is good (1083376/9.8 = 110549) but all the other ones are wrong. 

If we take the 2nd one and did the same calculation (1110109/9.8= 113276) and the software gave 111912. 

I don't know where this 111912 result is coming from. 

Here my injection list that shows how I performed my calibration.

Is it because my Cal.Std are not injected one immediately after another? 

I can see on my calibration curve that I has four point but the coefficient a given in the table is coming from the first injection only and not the average like I choose on the method process. 

We can't see that the a coeffiecient on the curbe at 110549 is coming from the first injection. 

So it seems that the Curve calculation is not with the average of the four peaks.

Why ?

thank you




Parents
  • Hello,

    Which injection did you use to show in the report for the calibration curve? Look at your calibration curve in data analysis and see what the A value is for that curve. Also note the way you are calibrating every sample or sample pair has a different calibration. Look at the example below where I have discreet brackets for the standards but are processing in the standard manner. You will note the A coff changes after each standard set. In older versions, before CDS 2.5 the coff would have changed after each standard injection not each bracket. In this case bracket is defined as cal std injections with no other sample types between them. If you want all your standard injections taken into account for all other injection types, you should use the overall bracket. 

    Marty 

  • Hi Martin. Thank you again for your precious help. 

    By changing to the "overall" bracketing mode, I finally had an average response based on the 4 TFT peaks founds in my 4 standard solutions (Cal.Std)

    My only concern it's that I can't show single response factor to each injection. Do you know how to do that? 

    It seems that only the average is showed. 

    But at least now, I can see that for TFT, 4 peaks are averaged in the calibration curve

    Thank you again for your help. 

  • Hi Martin the IS in our procedure is not for calculations but only for RRT since the Forensic analysis ask for it. 

    I only want to show RF in the single injection instead of showing the average response factore coming from my 4 standard solution injections. 

  • If you want to show the RF of an individual injection, you can't use the calibration curve average RF, which is what it seems you are referencing. I'm not sure if there is already a preset field for the RF of an individual injection, but the calculation is fairly simple to do anyway. In the RF column value expression just put =Area/Amount (where Area here is either Peak_Area or Compound_AreaSum depending on your case and Amount here is probably Compound_CalibAmount)

  • Hello,

    Just add a column to that table and use  =Peak_Area / Compound_CalibAmount as the expression for the default response per amount calibration. The yellow below is the CDS RF calculation and the blue is the per injection RF calculation value I created. 

  • Hi Martin. 

    In my example above, we see that the RF is the same for my 4 injections, like if the overall bracket mode is only show the average RF of the injections. 

    Do you know if I did something wrong or it's normal and I need to to the calculation like you just mentionned? 

  • Hello,

    That is what my example would have looked like with a single level standard. The previous example was a linear regression with 3 points not forced through the origin. One note when using overall bracket make sure all the injections that need the calibration applied are inside the standard bracket. 

  • perfect. It makes sense. Thank you for your help again. I will add the RF calculation like you have shown so our quality revisor will have access to the individual result instead of having the average RF. 

Reply Children
No Data
Was this helpful?