Hello,
What version of Chemstation are you using? The typical way to enable audit trails would be in the CS admin tool. You should be able to do it at the method level. Did you see the dialog to confirm the activation of the audit trail? Did you save the method after enabling the audit trail?
Marty
Yea I save method after enable audit trail
how can i obtain this window
Hello Mhawny,
The way you were enabling audit trails before is considered the manual way where individual methods are enabled by hand by the user. The way Marty describes is how to enforce audit trails to be enabled on all new methods automatically.
The screenshot Marty gave is the Chemstation Administration Tool and can be accessed from the Windows Start Menu -> Agilent Technologies -> Chemstation Administration Tool.
To be able to launch this tool the user account you are logging into Windows with needs to be in the Windows local group called CSAdministrators. After adding a Windows local user or domain account to the local Windows CSAdministrators group launch the Chemstation Administration tool and checkmark enable Method and Results audit trails to have the audit trails automatically generated with new methods and resuts.
-Bill
Hello Mhawny,
The way you were enabling audit trails before is considered the manual way where individual methods are enabled by hand by the user. The way Marty describes is how to enforce audit trails to be enabled on all new methods automatically.
The screenshot Marty gave is the Chemstation Administration Tool and can be accessed from the Windows Start Menu -> Agilent Technologies -> Chemstation Administration Tool.
To be able to launch this tool the user account you are logging into Windows with needs to be in the Windows local group called CSAdministrators. After adding a Windows local user or domain account to the local Windows CSAdministrators group launch the Chemstation Administration tool and checkmark enable Method and Results audit trails to have the audit trails automatically generated with new methods and resuts.