How do I include in OpenLab Software the criterias for system suitability including the pass and fail?
It is possible to configure it?
How do I include in OpenLab Software the criterias for system suitability including the pass and fail?
It is possible to configure it?
Hi
if you have in mind the evaluation of SS in report, I resolved by defining aggregator for resolution (AGGAvgResFac1) and group by compound name and created a text box with expression as follows
IIf(Prsd(AGGAvgResFac1(Compound_Name))>2.0,"FAILED","PASSED")
I hope it will work for you.
Mike
Hi
if you have in mind the evaluation of SS in report, I resolved by defining aggregator for resolution (AGGAvgResFac1) and group by compound name and created a text box with expression as follows
IIf(Prsd(AGGAvgResFac1(Compound_Name))>2.0,"FAILED","PASSED")
I hope it will work for you.
Mike