Openlab user updates impossible; online signals missing/frozen; some injections mysteriously won't calculate some compounds which are otherwise integrated properly.

Hi, I know you're not allowed to go to the doctor with multiple problems but here we go.

Updates are impossible.  I received a set of instructions with multiple options for (I'm assuming) multiple openlab variations which I could not tell the difference between.  Trying several of them, one provided me with around 15 updates that I "needed".  All but 1 or 2 came back with errors so did not actually update most of them.

The reason I was told to get updates and why I spent a considerable amount of company time to get internet and follow these unfriendly instructions was to avoid "disappearing chromatograms", see the post.  Did this update fix this?  How would I know? The problem only happened 3 times in a year, each devastating as you can imagine.  Something to do with a printer bug when you don't have a printer.

Now after that update my online signal is frozen everyday within a couple of minutes.  I don't know all day long what the machine is doing and just hope that it's running fine.  Also, many of the chromatograms won't calculate some of the compounds even though the peaks look integrated just like the rest of the chromatograms.  If my retention times match the standards why are the "amounts" coming up blank.  

Rebooting has not fixed either of these and some of my work and validations are on hold as a result.  Will call for help next week but I thought it's important to document my issues here.

Thanks and best of luck!

PJ

Parents
  • Hi PJ,

    What version of OpenLab do you have? ChemStation/CDS 2.x/EZChrom?

    With regards to updates. Each update does come with an extensive readme file that explains which update to apply to which type of installation. OpenLab CDS 2.x has standalone workstations through to client/server installations, and depending on which installation type you have depends on which specific bits of the update apply. If you are unsure of your installation type, you can find this in the Control Panel under Administration>System Configuration. If the Setup states "Workstation" then you have a standalone workstation and should only apply those updates. Updates are also cumulative. This means if update 3 has been installed and you install update 5, then only items for update 4 and 5 are added, as 1/2 and 3 already exist.

    As for the missing online signals. What type of instrument are you collecting the data from (LC/GC/LC-MS/GC-MS). How long are your runs and what is the online signal time range?

    With respect to calibration. Are all the compounds you are looking for 'identified' in the compound table, and do they have the correct calibration values in the compounds calibration table? And does the sequence have the calibration defined within it and with the correct levels? Load in some demo data from the install media (Setup>Tools>Support>Demo_Data) and see if that calculates values correctly. At least then you will know if it is the core software or something related to your data set

  • Hi, thanks for the response.  What version of openlab is not as easy a question as it seems.  Is it CDS 2.5? Is it M8410AA?  Is it "Workstation Plus"? Is it something else entirely? 

    I have an HPLC.  How long my runs are, online time range, all seems irrelevant (correct me if I'm wrong) as my problem is that the online signal is frozen almost immediately and stays frozen all day until the computer and/or HPLC are restarted.  I've covered the calibration issue above.  I can't wait for online signals again!  I'm calling Agilent and will report fully to the community.

    Thanks,

    PJ

  • Hi PJ,

    When you click "Run" the sequence list is meant disappear as it flips context to the running sequence. You just have to give it 10-20 seconds to hand off the sequence to the 'instrument control service'. Whilst it is doing that it shows a blank sequence table as it was before you entered anything. It is perfectly normal behaviour. Use the 'Run Queue' window to allow you to get back to that view once submitted.

    Have a look at the following YouTube clip (How to run a sequence in OpenLab CDS - YouTube). Navigate to the 18 seconds mark and then change the speed of the video to 0.25. Then play. You will see the sequence disappear after the 'Run' button is pressed and then a second later the running queue becomes visible. In actuality this can take 10 seconds or so depending on your PC hardware.

    Maybe try the OpenLab Help & Learning. Go to the Learning Modules>Sequences and have a look at the 'Submit and Modify a Sequence' section. This also shows how to get back to the information regarding what sample the sequence is on, what has been acquired and what is still to be acquired (Pending).

    As for the update error. I would register on the Laboratory Informatics support site (Agilent Software Service Desk) and log a software issue direct with them. Or, if you already have a Service Request, ask them to log a job on your behalf. They are specifically dedicated to the software, rather than the instrument hardware. They will request the log files etc and give you a place to out them. They may even do a remote session to see if they can find the issue, but this depends on your Software Maintenance Agreement status.

    Based on that fact you appear to have a Workstation Plus installation...I would also check you have the right privileges to do and see the things you are not seeing. There is one specific privilege 'Edit users own running sequence', which you should have, depending on your role. If not that could be why you cannot see what you have submitted.

  • Once again thanks for the rapid replies, it helps me to keep moving these issues towards resolution. 

    Now after a couple of weeks of posting this software issue to this forum I have now found out about a software support division which I never would have known about had it not been for this fantastic portal of knowledge. 

    The disappearing sequence does not re-appear like before my issues. That's what I stated above.  It's not a problem to address but merely a clue that might help with either the freezing online signal, disconnecting and failed sequences or my updating issues, my top Agilent issues at the moment.  I will reply if any of my issues are addressed so hopefully other hard working lab workers will not have as many mysterious issues.

    Thank you,

    PJ

  • Ok, the suggestion to address the disfunctioning Agilent software update process was to go to the software support website none of my 11 Agilent contacts told me about before.  Here's what I get when I want to "submit a request", a friendly reminder that technical questions will not be answered or responded to in the SMA Questions page.  Look at this screenshot and please tell me how I can submit a request to this Agilent website.  It looks like it's not possible nor is there any help anywhere here to be found on this "software support" website.  Pressing "here" for help above the word "Products" takes you to the SMA questions page not to any kind of help with issues.  Am I wrong or is this only for people with a service contract?  

    The other suggestion you made above, I have followed through on.  When I go to help and learning and even if I search for "modules", I get "no result".  Still wanting to try your suggestion somehow, I search for "Submit, modify"; also nothing.  What are you trying to show me here, please state it in this response if you're going to make one instead of sending me on a goose chase.  I already know I can see my position in "Run Queue".

    Thanks,

    PJ

  • FYI for anyone else being told to try the Agilent software support website, here's a lovely response completing the circle of futility and sending me back to the general phone line. Thanks a lot!

    "Hello

    This site is for client/server customers. I do not see a client/server system at Olli Brands Inc. For workstation support please call, 1-800-227-9770. Thank you!"

  • Hi PJ,

    That response is patently wrong. The Software support site is for any installation type, so long as you have an active Software Maintenance Agreement. This is different to the service contract.

    If you cannot find the modules section in the help and learning, that is a worry. If you have the install media, then you can just install the help and learning on a non instrument PC and have a look at it there. However, it should work on your instrument PC also. (A repair may be necessary)

    Now this is a big IF. If you have another PC/laptop, could you possibly install the software onto that, to see if the same thing occurs? You will get 60 days to test/check on the new PC and see if the software has the same problem.

    If you are having disconnection issues with the instrument I would check the network cabling from instrument to the switch and then to the PC. It may also be the network switch. I have seen some of the smaller plastic ones roll over and die mid sequence and never recover! It is also not unheard of for one specific port to die or become faulty, so changing the port the network cable is connected to might help. Bad network cables and 'iffy power can also be a cause of intermittent and hard to pin down disconnections. So if you know your power is good then look at the cabling. Go Cat 6 over 5e, just to be safe, But again I have seen dodgy Cat 6 cables cause problems too.

    It sounds counter intuitive to look at the hardware when you have a software issue. But as the software is completely dependent on the hardware platform of the instrument, PC and the network, I like to make sure it is ok. It eliminates one half of the equation so we can pinpoint our focus.

  • Thanks again for your thoughts on the computer disconnection from the LC.  I'm not sure what to do with that info though. Is there a way to test the cables or network switch?  

    "That response is patently wrong." 

    Are you saying that your own company does not produce the message below and that I'm wrong or that the message your company is providing customers is wrong?

    "Hello

    This site is for client/server customers. I do not see a client/server system at Olli Brands Inc. For workstation support please call, 1-800-227-9770. Thank you!" - Agilent website

    I see you're also trying to help me find "help and learning" topics by installing some software on another computer just to find advice that you want to point me to about the sequence table disappearing.  Let's park that suggestion as less urgent for now; if you'd like to share that advice directly, maybe I can try it. 

    Thanks,

    PJ

Reply
  • Thanks again for your thoughts on the computer disconnection from the LC.  I'm not sure what to do with that info though. Is there a way to test the cables or network switch?  

    "That response is patently wrong." 

    Are you saying that your own company does not produce the message below and that I'm wrong or that the message your company is providing customers is wrong?

    "Hello

    This site is for client/server customers. I do not see a client/server system at Olli Brands Inc. For workstation support please call, 1-800-227-9770. Thank you!" - Agilent website

    I see you're also trying to help me find "help and learning" topics by installing some software on another computer just to find advice that you want to point me to about the sequence table disappearing.  Let's park that suggestion as less urgent for now; if you'd like to share that advice directly, maybe I can try it. 

    Thanks,

    PJ

Children
  • Hi PJ,

    Whoever told you that the LI Support site is for client/server only, is wrong.

    As long as you have an active SMA (Software Maintenance Agreement) then any OpenLab CDS 2.x queries are allowed, even for workstation:

    System Configuration dropdown

    For the disconnections please also check to see if you have any Antivirus installed that might be interfering! Both Microsoft Endpoint and McAfee have been a little overzealous in their blocking of some services of late. So that might be an avenue to look at.

    The cables are trickier. Getting some new mid range price cables to test is how I would do it. or ask IT nicely if they have a couple of CAT 6 patch cables you can borrow to do some testing. Just make sure you return them! 

Was this helpful?