Macro for post-sequence reprocessing of data for OpenLab CDS Chemstation C.01.10

Hello,

My laboratory runs a multilevel calibration with internal standards on a headspace-GC-FID. Our sequence is set up to replace each level of the curve after each calibrator is analyzed and then calculate the concentration of the analyte with the calibrator based on the new curve. To ensure each calibrator is processed with the fully updated curve we always reprocess the data as soon as the sequence is finished. I know there's a Post-Run Command/macro field in the Sequence Parameters menu. Is there a macro or command that can be written to immediately reprocess all the data in a sequence once the sequence is finished running? I've tried some RunSequence commands, which work when you're in the Data Analysis view, however, since the sequence is running in the Method and Run Control view the RunSequence command attempts to reanalyze the sequence, which isn't what we want.

I'm new to OpenLab CDS Chemstation, so any help is appreciated!

Parents
  • Hello,

    Unless you are changing the calibration runs to samples or setting them to no update, simple reprocessing the result set as analyzed will not do what you are stating "To ensure each calibrator is processed with the fully updated curve we always reprocess the data as soon as the sequence is finished."

    Marty Adams

  • It works because by reprocessing the calibrators the software is replacing each calibrator with itself on the cal curve that was created by the initial sequence run. Since it's replacing itself the curve doesn't change and the concentration is calculated correctly. It's definitely a workaround, which is due to our unfamiliarity with the software. If there are suggestions for a more effective way I'm all ears.

  • Hello,

    Yes, if you only have single replicates of your standards that should work. The issue with trying to run a maco to reprocess the data set in method and run control at that point is the sequence has not been converted to reprocessing only at that point. This is why it is trying to re run the sequence. 

    Marty Adams

Reply
  • Hello,

    Yes, if you only have single replicates of your standards that should work. The issue with trying to run a maco to reprocess the data set in method and run control at that point is the sequence has not been converted to reprocessing only at that point. This is why it is trying to re run the sequence. 

    Marty Adams