Partial calibration in real time openlab CDS

(method partially calibrating on one column and not at all on the other)

Hey all,

I have not been able to figure out why I am getting a partial calibration in real time with my brand new method.  I am doing HS-GC-FID dual column analysis.

I have six analyte levels with an internal standard. 

I have properly populated the "Calibration" under my processing method with identifying my internal standards and associating my analytes of interest with their proper internal standard.  I put in the concentrations of all six levels.

Then, on my sequence I designate my calibration levels of 1-6 and use the "Clear all calibrations" from the dropdown for the first cal.std type so that a new curve should be generated.  However, I am only getting a two point calibration (my two highest standards) on only one of my columns.  The other column's curve is totally blank. 

All the peaks/data are present in all six calibrators and when I reprocess that sequence with the method I used in the sequence in the first place, it properly calibrates and all six points are present on both columns!  How is this not working in real time, but its working in reprocessing?

(note successful 6-point calibration from reprocessing of the exact same method)

Parents
  • Hello,

    After you setup the method in your first result set did you use the update master method selection? When you link a method to a result set it copies that method into the result set and any changes you make are on that method copy in the result set. If you want to update the original master method to use when running sample then you need to use the update master method.

    Marty Adams

  • Ok, I'm still doing something wrong (and I can't add pictures like I could yesterday for some reason)

    I Updated the Master Method as shown, and assumed it updating it would save it.  I ran another batch and got the same two point calibration as before. 

    So today I reprocessed that data to get all six calibration points, updated the master method, then SAVED Method (which I hadn't done before because I thought it would automatically save). 

    But when I exited to go back to the sequence, I got the message "Unsaved Changes--There are unsaved results-- Do you want to proceed closing without saving?"  I couldn't find where I could save anything else because "Save Method" was already greyed out since I had just done that.

    I re-ran the sequence and I am STILL getting the 2-point calibration on one column and nothing on the other.

    Do I have to do something to link the method?  I don't understand the link icon.  I tried that, but it didn't work, either.

  • The "Unsaved changes" message refers to the results not the method.  When you reprocess data in Data Analysis, you need to save the results before you close the program, or before you generate reports.

    Back to the cal. curve problem:

    Have a look at the Injection List in Data Analysis.  Are all of your standards correctly identified and is Clear All calibrations set just for the first standard?

    Does the processing method still have 6 levels.  Check this on the General tab of the Compounds...Calibration part of the method.

    Not sure why you can't share more images.  I think I saw that another user had the same problem today.

    /Andy

Reply
  • The "Unsaved changes" message refers to the results not the method.  When you reprocess data in Data Analysis, you need to save the results before you close the program, or before you generate reports.

    Back to the cal. curve problem:

    Have a look at the Injection List in Data Analysis.  Are all of your standards correctly identified and is Clear All calibrations set just for the first standard?

    Does the processing method still have 6 levels.  Check this on the General tab of the Compounds...Calibration part of the method.

    Not sure why you can't share more images.  I think I saw that another user had the same problem today.

    /Andy

Children
  • Ok, thanks for the information on the results.  I don't care about saving reprocessed results since my focus is to just get a method established that works the way I need it to.  As long as I don't need to do this for updating my master sequence to work.....(Looks like pic insertion is now working)

    In my Injection List in Data Analysis, my standards are correctly identified, levels designated, and "clear all calibrations" is set just for the first standard:

    My processing method still has all 6 levels:

    I also have internal standards correctly designated along with the correct association of analytes:

     

Was this helpful?