How to create a formula with calculations of custom fields values in OpenLab CDS

Hi 

 

am developing an intelligent report for quantification in OpenLAB CDS 2.4 and I want to include an acceptance criteria in my results table. The acceptance criteria will pass if the compound amount is within +/- 10% of a reference value.

 

I was able to do so when only one analyte is quantified (typing the reference value in  the sample amount section) using the following expression:

=Iif(Compound_Amount<=(Sample_Amount-(Sample_Amount*0.1)), Compound_Amount>=(Sample_Amount+(Sample_Amount*0.1)),"Pass"

 

However, I haven't been able to create a formula for when several analytes are quantified. I've created a compound custom parameter "Expected amount" to be able to give reference values to each of the quantifiable analytes and I have been able to use the expression to calculate accuracy but I need help creating the formula for the acceptance criteria.

 

=Iif(Compound_Amount<=(CFE((Compound_CustomFields),"ExpectAmt")-(0.1*CFE((Compound_CustomFields),"ExpectAmt"))), Compound_Amount>=(CFE((Compound_CustomFields),"ExpectAmt")+(0.1*CFE((Compound_CustomFields),"ExpectAmt"))),"Pass")

 

Thank you for your time

Parents
  • Hello,

     

    I have added tags to this post to increase visibility.

     

    If you want the expression to output "Pass" if 0.9*Sample_Amount<Compound_Amount<1.1*Sample_Amount and output "Fail" otherwise, then you will want an expression like below:

     

    =IIf(Compound_Amount<=(Sample_Amount-(Sample_Amount*0.1)) OR Compound_Amount>=(Sample_Amount+(Sample_Amount*0.1)),"Fail","Pass")

     

    A similar expression can be used with your custom parameter. 

Reply
  • Hello,

     

    I have added tags to this post to increase visibility.

     

    If you want the expression to output "Pass" if 0.9*Sample_Amount<Compound_Amount<1.1*Sample_Amount and output "Fail" otherwise, then you will want an expression like below:

     

    =IIf(Compound_Amount<=(Sample_Amount-(Sample_Amount*0.1)) OR Compound_Amount>=(Sample_Amount+(Sample_Amount*0.1)),"Fail","Pass")

     

    A similar expression can be used with your custom parameter. 