Compare Open Lab 2.1 / 2.2 / 2.3

Hello,

 

We are planning to move our Open Lab 2.1 to 2.3

Im not sure if there is a documentation which compare 2.1 to 2.3.

So I need the Step from 2.1 to 2.2 too.

 

 

So I would be happy to have them both.

 

Important question.

Can we migrate the data, we already created ?

Can we use the same Reports ( rdl files ) ?

 

Can the service technician just upgrade it, or do he need to reinstall everything.

Is it a clean installation ( wipe everything ) ?

 

Regards

Parents
  • Hello mr.white,

     

    I have attached the release notes for OpenLab CDS 2.3. This document outlines the major feature modifications introduced with each release of OpenLab CDS 2.x from 2.1 to 2.3.

  • I forgot to address your other questions. In-place upgrades from a OpenLab CDS v2.1 WorkStation Plus to a v2.3 WorkStation Plus are supported, meaning that your 2.1 software does not need to be removed in order to install 2.3. The data, methods, reports, etc... should transfer without issue.

  • Hello Ryoboyle,

     

    thank you for the fast reply *insert medal emoticon*

     

    Ok since we currently using the Content Manager we need Workstation plus.

    I just check the pdf File and the new features look very good.

     

    Especially:

     

    Loading methods from another project

     

     

    Allows authorized users to browse to another project, open a method and save a copy of that method to their current project or link it to a result set.

     

    Overlay chromatograms from different projects

     

     

    Allows users to load data from other projects to compare results.

     

     

    Zoom in/out of report preview with mouse wheel.

    1.) During the run ? I recognized that it was missing in 2.1

     

     

     

    Use of complex custom fields in filter, grouping and sorting expressions.

    2.) Can we get a pdf File, for the use of it, when we purchase the Software and the qualification service  ?

     

     

    Blank subtraction

     

     

    Use the blank subtraction to calculate a clean chromatogram with contribution of the analytes only.

    3.) Does it substract the Injection Peak as well ?

     

     

    4.) XML Support. Do this Version have XML Support now ? Like the good old chemstation.

     

     

    Thank you very much. If you reply please refer to the numbers.

     

    Regards

     

     

     

     

Reply
  • Hello Ryoboyle,

     

    thank you for the fast reply *insert medal emoticon*

     

    Ok since we currently using the Content Manager we need Workstation plus.

    I just check the pdf File and the new features look very good.

     

    Especially:

     

    Loading methods from another project

     

     

    Allows authorized users to browse to another project, open a method and save a copy of that method to their current project or link it to a result set.

     

    Overlay chromatograms from different projects

     

     

    Allows users to load data from other projects to compare results.

     

     

    Zoom in/out of report preview with mouse wheel.

    1.) During the run ? I recognized that it was missing in 2.1

     

     

     

    Use of complex custom fields in filter, grouping and sorting expressions.

    2.) Can we get a pdf File, for the use of it, when we purchase the Software and the qualification service  ?

     

     

    Blank subtraction

     

     

    Use the blank subtraction to calculate a clean chromatogram with contribution of the analytes only.

    3.) Does it substract the Injection Peak as well ?

     

     

    4.) XML Support. Do this Version have XML Support now ? Like the good old chemstation.

     

     

    Thank you very much. If you reply please refer to the numbers.

     

    Regards

     

     

     

     