Good Morning,
Would you please advise what are the possible causes of getting a low energy value of 16 on my cary 630?
Good Morning,
Would you please advise what are the possible causes of getting a low energy value of 16 on my cary 630?
Hi Serge, thanks for your question. You should change the source assembly (reference G8043-67400 purge fiitting included in the reference ), after replacing the source you should purge with N2.
Thanks
Regards
Hi Serge, thanks for your question. You should change the source assembly (reference G8043-67400 purge fiitting included in the reference ), after replacing the source you should purge with N2.
Thanks
Regards
Thank You FCA, Appreciated.