I have issue withe 8453 power switch led on , front panel led on (red light)
after verification the both lamps are not on..
how i can check the sls board or what i can check ?
THANKS I
I have issue withe 8453 power switch led on , front panel led on (red light)
after verification the both lamps are not on..
how i can check the sls board or what i can check ?
THANKS I
Hi
First replace lamp then check...
both lamps are new, the problem appeared after replacing the lamps and dusting the instrument
Lamps are litting or not..If litting then alignment issue and if not litting then electronics issue...
how i check the litting lamp and i proceed of alignment
When you power on the instrument then after sometime did u see any light from lamp compartment...
When you power on the instrument then after sometime did u see any light from lamp compartment...
no any light,
Then i have a doubt that may be lamp power supply source is not okay.
If you can help from below documents then okay, otherwise, i will suggest to call your local agilent support to check it onsite.
Again check all the connector of lamps..