Hi Everyone,
Due to random readings of my samples, i ran wavelength and photometric accuracy test. Both of them fails, do i need to replace the lamp or is it because of the wavelength drive gear?
thanks.
Hi Everyone,
Due to random readings of my samples, i ran wavelength and photometric accuracy test. Both of them fails, do i need to replace the lamp or is it because of the wavelength drive gear?
thanks.
Hi
Check above link.
If lamp is very old then may lamp problem otherwise hardware issue...
Hi,
lamp age is around 10 years. but, i do not know the exact data. i will check soon. thank you.
Hi..
could be issue with Lamp, if old. If not then could be source alignment issue as well.
it fails only at 360.900 line not found FAILED
Tolerance for 360.900 nm ±1.00 nm
line scan test %T around 10. photometric accuracy test all steps fail. One of the sample readings are;
Sample 1 -0.1619
Sample 2 -0.0923
Sample 3 -0.1376
Sample 4 -0.0444
Sample 5 -0.0011
Sample 6 0.0038
Sample 7 -0.0828
Sample 8 0.0145
Sample 9 -0.1776
Sample 10 -0.0271