Hi Everyone,
Due to random readings of my samples, i ran wavelength and photometric accuracy test. Both of them fails, do i need to replace the lamp or is it because of the wavelength drive gear?
thanks.
Hi Everyone,
Due to random readings of my samples, i ran wavelength and photometric accuracy test. Both of them fails, do i need to replace the lamp or is it because of the wavelength drive gear?
thanks.
Hi,
lamp age is around 10 years. but, i do not know the exact data. i will check soon. thank you.
it fails only at 360.900 line not found FAILED
Tolerance for 360.900 nm ±1.00 nm
line scan test %T around 10. photometric accuracy test all steps fail. One of the sample readings are;
Sample 1 -0.1619
Sample 2 -0.0923
Sample 3 -0.1376
Sample 4 -0.0444
Sample 5 -0.0011
Sample 6 0.0038
Sample 7 -0.0828
Sample 8 0.0145
Sample 9 -0.1776
Sample 10 -0.0271