Cary 50

Good morning, I'm writing about an error when I carry out a kinetic, for example with 7 cuvettes: every time the instruments reads the cell, I read on display:

<missing string resource 14736>1

and so on for 2, 3, 4, 5, 6, 7.

What is the reason?

It seems that the word "sample" is missing.

Thank you very much in advance.

  • Hallo Crici,

    seems as your software  was installed under local regional settings. Please check under regional settings that your settings in teh control panel look like this:

    ..you can get access under control panel->Region->Adminsitrative to get this information. If you have a mismatch here, change the settings, uninstall and reisntall the software. That should fix it!

    Kind regards,

    Frank.

Was this helpful?