6300 ion trap how to adjust capacity of RF-Generator

[V] RF ramp set voltage AI15   passed

[V[ RF ramp actual voltage AI31  0.01   -0.05 0.05   passed

                                                     0.01      4.90 5.10  failed

                                                     0.01      8.90 9.10   failed

[A] RF ramp generator current AI39   -0.00       -0.05  0.20  passed

                                                            0.00         0.08  0.60  failed

                                                            -0.00        0.02  1.10  failed

[V] RF ramp check voltage AI62          4.38          4.00  4.75  passed

                                                             4.38          4.50  5.50   failed

                                                              4.38        5.00  7.00    failed

 

CTS RF overload after ramp OVLS       5              4     failed

 

everything else passed 