Good Morning. I have a G7129A Vialsampler with error 34331, transport arm position lost during x axis movement. Have any engineers successfully rectified this by adjusting belt tension or is it better to change the transport assembly (G7129-60600)?
Good Morning. I have a G7129A Vialsampler with error 34331, transport arm position lost during x axis movement. Have any engineers successfully rectified this by adjusting belt tension or is it better to change the transport assembly (G7129-60600)?