Hello i asked earlier about HED Fault i encounterd
The Autotune gives error
Filament turns off
I check for air leak ... the door closes tightly immediately when the vaccum pump turns on
Hi Vac is N/C
And this is the error
Hello i asked earlier about HED Fault i encounterd
The Autotune gives error
Filament turns off
I check for air leak ... the door closes tightly immediately when the vaccum pump turns on
Hi Vac is N/C
And this is the error
I have already cleaned detector and EM. But still no change.
There could be issues with either control signals to the HED power supply or the power supplied to the HED power supply. These electronic issues are down in the depths of the instrument and should be handled by a trained service engineer. Please contact your local country's Agilent service.