Hi,
Which is the correct flow in saver gas for GC/MS? I have it at 20ml/min.
Hi,
Which is the correct flow in saver gas for GC/MS? I have it at 20ml/min.
You can swap out cylinders to do it, but that means a lot more tubing, etc. to purge when you want to run. With the EPC unit, you would simply load the appropriate method to use it and the switch is done at the instrument. Much less tubing to purge and quicker to have the system ready to run when you need it.
If you're idling the system for a few weeks at a time, it may be simpler to shut the system down. More details on how to safely do that can be found at How to shut down an Agilent GC/MS for power fails up to long term storage 17Mar2020.pdf
We have 13 GCs on 5-17 mL total flow when idle and they work just fine for environmental samples and surge events the CDC sends us. 552.3 instrument probably has 3500 injections with no maintenance and a 17 month old curve. Some mass specs with a concentrator for 524 or TO-15 will see water climb in sleep mode but it all blows out after the tune is off. If there is column damage it's still cheaper to toss a new one in than foot the helium bill in my opinion, I haven't seen abnormal wear yet 9 months in. Our CDC unit has been running like this for years.