I am running GC 6890 (FID detector) alongside G1888 headspace.
I had noticed 4 of my later eluting peaks got increasingly worse in terms of shape and reproducibility. After running blanks (no injection and also blank vials), it seemed like the cause of this turns out to be carryover. The more blank vials I run after my highest standard, the smaller the carryover peaks become. However, it is not time efficient for me to have to run 3 blanks in between standards in order to get decent peak shapes.
GC Paramaters
OVEN
Initial: 60C (hold 0.1min) ramp to 300C (12.5C/min), hold 6 min
Carrier gas: Helium
I am using Rxi-624Sil MS column.
I have attached chromatograms to illustrate my problem.
I am relatively new to GC so please do not hesitate to suggest something obvious.
Thanks!
Edit to add more instrument parameters:
Headspace
G1888 Headspace
Configuration
Headspace Device: Agilent G1888 Headspace Sampler(g1888)
Communications Mode: 10.1.1.102
Serial Number: IT00739016
Vial Size (mL): 20
Handshake Mode: Wait if GC Not Ready
Oven Stabilization Time (min): 1.00
Pressure Units:: psi
Carrier Connection: Front Inlet
Vial EPC: None
Loop Size (mL): 1
Method
GC Cycle Time (min): 40.00
Inject Time (min): 1.00
Loop Equilibration Time (min): 0.05
Loop Fill Time (min): 0.50
Loop Temperature (°C): 135
Oven Temperature (°C): 120
Shake: Off
Transfer Line Temperature (°C): 140
Vial Equilibration Time (min): 30.00
Vial Pressurization Time (min): 0.08
Vial Pressure (psi): 0.00
Carrier Gas Pressure (psi): 0.00
GC
OVEN
Initial: 60C (hold 0.1min) ramp to 300C (12.5C/min), hold 6 min
Carrier gas: Helium
FID
Temp: 320C
Hydrogen flow: 30mL/min
Air flow: 450mL/min
Helim: 45mL/min