BIO-DIS Reciprocating Cylinder Apparatus, 300 mL G7970A , samples probes

We recently have installed BIO-DIS Reciprocating Cylinder Apparatus, 300 mL G7970A and 850-DS Dissolution SamplingStation. The sample probes that collect the sample from vessels come in a lenthg that serves only dissolution medium volume of 300 ml. But we have an analytical method which requires dissolution medium of 150 ml. S o my answer is , if there is a longer sample probe for BIO-DIS Reciprocating Cylinder Apparatus, 300 mL G7970A?

Thank you in advance

  • We do not have sample probes that will allow for automated sampling at 150 mL in the 300 mL vessel.

    150 mL in a 300 mL vessel is not an ideal method.

    You would need to manually sample at this volume since it is beyond the operational minimum volume.