Is there a tubing diagram available for an 708DS -> 806 Syringe Pump -> 8000 Sampling Station? Also what is the recommended tubing for each connection step? Thanks.
Is there a tubing diagram available for an 708DS -> 806 Syringe Pump -> 8000 Sampling Station? Also what is the recommended tubing for each connection step? Thanks.
We do not have a diagram for the tubing connections.
The part numbers for tubing will depend on your instrument configuration.
Below are links to the 8000 operator's manual and the dissolution source book.
https://www.agilent.com/cs/library/usermanuals/public/Agilent_8000_Op_Man.pdf
We do not have a diagram for the tubing connections.
The part numbers for tubing will depend on your instrument configuration.
Below are links to the 8000 operator's manual and the dissolution source book.
https://www.agilent.com/cs/library/usermanuals/public/Agilent_8000_Op_Man.pdf
HI Thanks for your response.
we have a 780DS, a 806 Syringe Pump and a 8000Sampling Station (with Internal Media Replacement Pump)
Do we need the Syringe Changer to complete the setup or can we go ahead with what we have and configure the autosample to bath setup?
I do not know what you mean by syringe changer to complete the setup.
The 8000 will work with the 806 syringe pump and dissolution apparatus.
Please see the 8000 operator's manual for the cable connection information.
Sorry i meant filter changer.
A filter changer is not required.