Is there a way to add different sequential start time options for the analyst to chose from when running a method. Or a way for the analyst to specify the sequential start timing without giving them access to modifying methods?

I am looking for options since different analysts might be quicker or slower when sampling and it will be good if each could choose their own sequential dropping time based on the time they need for sampling each vessel, without having to give access to modifying/creating/deleting methods. We are using Dissolution Workstation software V 4.1.6.28655.

  • The Sequential start parameter (time between dosage introduction) is set in the method.

    Access to change this value cannot be given to users without Method editing rights.

     

    The operator can choose a sequential start when launching the method but they cannot adjust the time between DDM openings at that point. 

Was this helpful?