Packed a 5190-2295 inlet liner with wool

Anyone packed a 5190-2295 inlet liner with wool without having problems with the wool not staying packed and shifting during a few runs?

Parents
  • You don't repack glass wool into a liner as this will influence the flow restriction of it. More densely packed glass wool gives higher resistance, less densely packed and it shifts around. Also Ultra Inert liners are packed with Ultra Inert glass wool and you don't want to break the glass fibers by self-packing as those edges and surfaces are not ultra inert any more.

    It is better to use a new liner instead.

    Regards,

    Norbert

Reply
  • You don't repack glass wool into a liner as this will influence the flow restriction of it. More densely packed glass wool gives higher resistance, less densely packed and it shifts around. Also Ultra Inert liners are packed with Ultra Inert glass wool and you don't want to break the glass fibers by self-packing as those edges and surfaces are not ultra inert any more.