Hi,
short question - please.
my column for determination "hydrogen" is starting bleeding, what can you do to eliminate this process of bleeding?
thanks for providing me feedback,
Hi,
short question - please.
my column for determination "hydrogen" is starting bleeding, what can you do to eliminate this process of bleeding?
thanks for providing me feedback,
dear,
we use three types of different columns (molsieve 5A, HP-AL/s, HayesepQ).
kind regards,
Hi, this is still not sufficient.
I assume you use a column switching system with HayeQ as first column, switching the first eluting peak to M5A for the permanent gases, and backflush the HayeQ to Alox for hydrocarbon analyses?
Molsieve and aluminium oxide are inorganic compounds and normally cannot bleed, HayeQ might be bleeding if the temperature was above 250 °C.
As you are in Belgium I suggest calling your nearest Agilent office and connect to the CSD Technical Support team as this is no simple question.
Regards,
Norbert
Hi, this is still not sufficient.
I assume you use a column switching system with HayeQ as first column, switching the first eluting peak to M5A for the permanent gases, and backflush the HayeQ to Alox for hydrocarbon analyses?
Molsieve and aluminium oxide are inorganic compounds and normally cannot bleed, HayeQ might be bleeding if the temperature was above 250 °C.
As you are in Belgium I suggest calling your nearest Agilent office and connect to the CSD Technical Support team as this is no simple question.
Regards,
Norbert