HP-INNOWax 19091N-133I

HP-INNOWax column 19091N-133I cannot be selected during method development, is it possible to select 19091N-133ILTM (ILTM format) as an alternation choice?

Parents Reply
  • Dear Dr. No,

    Thanks for your reply.

    The maximum temperature allowed for part number 19091N-133ie is 445℃ and the one we concerned, part number 19091N-133i, is only 270℃. If we select -133ie in the GC method, the program will keep on process even somehow above 270℃ without warning. Hope you suggest how to implement column 19091N-133i if it is not a choice in Column Selection in Method Setting. 

    Best regards,

    Gordon

Children
Was this helpful?