HPLC COLUMN PROTECTION
To save the analytical column Reverse Phase ZORBAX Eclipse Plus C18 (100 x 4,6 mm, 3,5 μ), the use of Javelin Colfilters (1 x 10 pcs.) cod. S90199511 is obligatory.
HPLC COLUMN CONDITIONING
Install a new analytical C18 Reversed-Phase column ZORBAX Eclipse Plus C18 (100 x 4,6 mm, 3,5 μ) .
Disconnect the detector and filter 30 ml of H2O : Acetonitrile (20 : 80 v / v) set flow of 1,0 ml/minute. Don’t recycle the washing solutions. Flux 30 ml of a solution of H2O : Acetonitrile (80 : 20 v / v) at flow of 1,0 ml/minute. Don’t recycle the washing solutions. Condition the column with the mobile phase at a flow of 1,0 ml/minute and discharge the first 20 ml. Finally inject the Chemical Standard and verify the quality of the HPLC run.
It’s NOT possible to make analysis at recycling phase. If room temperature is 20 °C store the Mobile Phase at 2-8 °C between an analytical session and another.
COLUMN CLEANING
Disconnect the detector. Flux 15 ml of H2O : Acetonitrile or Methanol (80 : 20 v / v) set flow of 1,0 ml/minute and discharge. Flux a solution of H2O : Acetonitrile or Methanol (30 : 70 v / v) for 30 minutes discharge it. It is recommended to perform the washing of Analytical Column at the end of each analytical run. Store the column in a solution of H2O : Acetonitrile or Methanol (30 : 70 v / v).
HPLC COLUMN PROTECTION
To save the analytical column Reverse Phase ZORBAX Eclipse Plus C18 (100 x 4,6 mm, 3,5 μ), the use of Javelin Colfilters (1 x 10 pcs.) cod. S90199511 is obligatory.
HPLC COLUMN CONDITIONING
Install a new analytical C18 Reversed-Phase column ZORBAX Eclipse Plus C18 (100 x 4,6 mm, 3,5 μ) .
Disconnect the detector and filter 30 ml of H2O : Acetonitrile (20 : 80 v / v) set flow of 1,0 ml/minute. Don’t recycle the washing solutions. Flux 30 ml of a solution of H2O : Acetonitrile (80 : 20 v / v) at flow of 1,0 ml/minute. Don’t recycle the washing solutions. Condition the column with the mobile phase at a flow of 1,0 ml/minute and discharge the first 20 ml. Finally inject the Chemical Standard and verify the quality of the HPLC run.
It’s NOT possible to make analysis at recycling phase. If room temperature is 20 °C store the Mobile Phase at 2-8 °C between an analytical session and another.
COLUMN CLEANING
Disconnect the detector. Flux 15 ml of H2O : Acetonitrile or Methanol (80 : 20 v / v) set flow of 1,0 ml/minute and discharge. Flux a solution of H2O : Acetonitrile or Methanol (30 : 70 v / v) for 30 minutes discharge it. It is recommended to perform the washing of Analytical Column at the end of each analytical run. Store the column in a solution of H2O : Acetonitrile or Methanol (30 : 70 v / v).
You have the conditioning and cleaning procedures for your specific column in the manual delivered with the column.
Regards,
Norbert