C2C12 Cell detachment from 96 well plate

Hi, I am growing C2C12 cells in Seahorse 96 well plate for mitostress analysis after differentiation, but I have observed multiple times that the cells keep detaching, laminin coating also doesn't improve the condition. Usually the detachment takes place right before the analysis. I have tried modifying the seeding density multiple times but It doesn't seem to improve. Can you help me?

Parents Reply
  • Hi Pillai,

    Thank you for the additional clarification. I found this differentiation protocol using Matrigel, just in case you wanted to take a look at it. I have a few follow up questions:

    1. Can you tell me how many hours after differentiation the cells are being washed with the Seahorse media?

    2. How many cells/well are seeded?

    3. When washing, are you adding the media to the side of well rather than directly on top of the monolayer?

    4. How much media (uL) are you leaving behind when washing the plate?

    Thank you,

    Courtney

Children
Was this helpful?