I am planning to isolate monocytes for Seahorse analysis and wondering wich isolation method to use,
and is the purity of the isolated monocytes of importance when doing Seahorse ?
I am planning to isolate monocytes for Seahorse analysis and wondering wich isolation method to use,
and is the purity of the isolated monocytes of importance when doing Seahorse ?
Hi Imr,
Thank you for your question. The first thing I want to point out is that you want to make sure your cells are adhered to the bottom of the well. I am not sure which type of instrument you have, but below are our protocols for adhering non-adherent cells via Cell-Tak:
You do want them to be as pure as possible because any other cells that are seeded in the plate will contribute to your rates. We do not have an official recommendation as far as purifying methods, but you can use our Cell Reference Database to see what other users have done to purify. You can change the “Cell Type” drop-down to monocytes – there are many different papers listed that should be helpful!
Here are some papers I was able to find on there that I thought you might find interesting:
Additionally, here is a Jove Video specific to leukocytes that you may find helpful.
Please let me know if you have any other questions after taking a look at these documents/links.
Thanks!
Courtney
Hi Imr,
Thank you for your question. The first thing I want to point out is that you want to make sure your cells are adhered to the bottom of the well. I am not sure which type of instrument you have, but below are our protocols for adhering non-adherent cells via Cell-Tak:
You do want them to be as pure as possible because any other cells that are seeded in the plate will contribute to your rates. We do not have an official recommendation as far as purifying methods, but you can use our Cell Reference Database to see what other users have done to purify. You can change the “Cell Type” drop-down to monocytes – there are many different papers listed that should be helpful!
Here are some papers I was able to find on there that I thought you might find interesting:
Additionally, here is a Jove Video specific to leukocytes that you may find helpful.
Please let me know if you have any other questions after taking a look at these documents/links.
Thanks!
Courtney
Immobilization of Non-Adherent Cells with Cell-Tak for Assay on the Agilent Seahorse XFe/XF96 or XFp Analyzer
I have been following this protocol for cell attachment but the cells still couldn't attached. In the technical note, sodium hydroxide was mentioned in the Materials, but the use of sodium hydroxide was not mentioned in the protocol steps. The protocol in Cell-Tak is a little different. I am a bit confused. Please help.
Hi there,
Thank you for contacting Cell Analysis Technical Support. The protocol you mentioned does reference the manufacturer of Cell Tak Basic Protocol (see screenshot below). This is where the sodium hydroxide is used to neutralize the Cell Tak for the adhesive function to be activated. Have you gone to that referenced protocol as well? You will see it there. Please let me know if this helps or if you have any additional questions once you take a look.
Thank you,
Courtney