Hello,
As Cole said you need to be selecting the correct sample types in you sequence to use this. If you are doing that, the filter would be Sample_Type <> 1 to remove the calibration standards or Sample_Type = 3 for just samples if you have other…
Hello,
When you say results does that mean raw data or reports? For the raw data you can control the storage location using projects and then subfolders under the project. If you want the reports to go to a central share in a specific folder that…
Thank you Sabbas for your question.
The Tutorials and Help in Cary WinUV 5.3 is also relevant for many questions, applications and workflows a Cary WinUV 5.1.3 user might have. However, some sections might be specific and different for Cary WinUV 5…
Delayed solution here. But I was able to solve this problem by switching from glass wool liners to glass bead liners. The glass bead liners have more surface area. I assume this issue was exacerbated by contamination but it seems it was related to the…
Collection of Resources
Link to User Forum
View Agilent User Manuals, Application Notes, White Papers and Tools.
Join the Community and participate in the User Forum ! Find answers to questions, ask questions and share your insights…
B.07.05.2479 was released in September 2016. That version of Acquistion will work on WIndows10 64 bit but the MassHunter Qual and Quant from that time will not. The current revision of Acquisition tested on Windows 11 is 10.2 SR3. Upgrading software will…
This Information Applies To: All Agilent Gas Chromatographs, 6890, 6850, 7890, 7820, 8890, and 8860.
Issue:
A quick reference guide on the capillary column installation for the GC.
Inlets
Detectors
Split/splitless inlet Purged...
How to Add a GC Instrument in Agilent OpenLab CDS (Ver 2.x)
Multiple Results
This Information Applies To: Agilent OpenLab CDS (Ver 2.x)
Content:
The following procedure shows how to add a GC instrument in OpenLab CDS (Ver 2.x).
Steps to Follow:
1. Open OpenLab CDS Control Panel, in the Instruments interface, select...
This Information Applies To: Agilent Intuvo 9000, 8860, 8890, 7890A/B, 7820A, 6890, 6850 GC Systems with Split/Splitless, or Multimode Inlets.
Issue:
This article describes how to replace the inlet septum for Split/Splitless and multimode inlets...
This Information Applies To: Agilent GC Systems
Issue
The data rate determines the number of sampling points across the chromatographic peak and the sample background. The data rate is based on the width of the analyte speak (in minutes) and the...
This Information Applies To: Agilent GC Systems
Issue
Excessive baseline noise can be introduced at different points including sample preparation, sample introduction, separation, and the detector. The excessive background can cause reduced signal...
This Information Applies To: Agilent GC Systems
Issue
Baseline instability that is not reproducible across multiple runs can be introduced at different points including sample introduction, separation, and the detector. The baseline instability...
This Information Applies To: Agilent GC Systems
Issue
Peak fronting is present in a chromatography peak when it has an excessive asymmetry with a leading edge. A normal peak is almost symmetrical.
Other chromatographic problems are identified...
This Information Applies To: Agilent GC Systems
Issue
A shift in retention time can occur for single, multiple or all peaks.
Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument...
This Information Applies To: Agilent GC Systems
Issue
A change in peak size can occur for single, multiple or all peaks.
Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument...
This Information Applies To: Agilent GC Systems
Issue
Peak splitting is when a single compound is shown in the chromatogram as two peaks usually unresolved, or one of the peaks is a shoulder.
Other chromatographic problems are identified in Basic...