Hello jeremy_lai ,
Available graphics options for reports can be found on the Graphics settings tab in the Report Method editor. From what I can find, the Qualifier graphic does not have any specific display options available. The only ones that apply…
Hello bertinettis ,
I apologize for the delayed response.
There is an option in quant to enable pre-set peak annotations in the Add-Ins from the Tools menu/ribbon.
These will then appear in the Compound Information window. Once a peak has been…
The restriction capillary is put in place of the column, to protect the column and run a wear in procedure after replacing the pump seals. Or to flush the system during tests. Do you still need assistance with your request?
luciayuan Why DMF cannot go through channel D? Which pump do you have? Isocratic or Quaternary pump? For the leak rate test, I would suggest pressurize the pump at the purge valve to find out where it is leaking. I would suggest log a ticket with your…
This Information Applies To: Agilent GC Systems
Issue
Excessive baseline noise can be introduced at different points including sample preparation, sample introduction, separation, and the detector. The excessive background can cause reduced signal...
This Information Applies To: Agilent GC Systems
Issue
Baseline instability that is not reproducible across multiple runs can be introduced at different points including sample introduction, separation, and the detector. The baseline instability...
This Information Applies To: Agilent GC Systems
Issue
Peak fronting is present in a chromatography peak when it has an excessive asymmetry with a leading edge. A normal peak is almost symmetrical.
Other chromatographic problems are identified...
This Information Applies To: Agilent GC Systems
Issue
A shift in retention time can occur for single, multiple or all peaks.
Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument...
This Information Applies To: Agilent GC Systems
Issue
A change in peak size can occur for single, multiple or all peaks.
Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument...
This Information Applies To: Agilent GC Systems
Issue
Peak splitting is when a single compound is shown in the chromatogram as two peaks usually unresolved, or one of the peaks is a shoulder.
Other chromatographic problems are identified in Basic...
This Information Applies To: Agilent GC Systems
Issue
This test isolates the point of contamination to the sample introduction system, which includes the gas supply, gas supply lines, and the sample inlet. It does not include the syringe or autosampler...
This Information Applies To: Agilent GC systems
Issue
Peak tailing is present in a chromatography peak when it has an excessive asymmetry with a trailing edge. A normal peak is almost symmetrical.
Other chromatographic problems are identified...
T his Information Applies To: Agilent 8890/8860/7890/7820/6850/6890 GC Systems with TCD (Thermal Conductivity Detector)
Issue
Problems observed with thermal conductivity detector (TCD) results can have different underlying causes.
Resolution...
This Information Applies To: All GCs, 6890, 6850, 7890, 7820, 8890/8860, Intuvo 9000
Issue:
This article is a troubleshooting guide for when the Flame Ionization Detector (FID) does not ignite or the flame keeps going out.
Cause:
If the FID...