This Information Applies To: Agilent AriaMx real-time PCR system, AriaMx real-time PCR optical modules
Issue
The AriaMx has the capacity to accommodate six optical modules, and the scanning optics configuration ensures distinction between fluorescent dyes and individual samples. Agilent has not...
This Information Applies To: Agilent AriaMx real-time PCR system
Issue
Can I use reaction volumes larger than 30 μL on the AriaMx?
Resolution
There is no problem with using higher reaction volumes, but the instrument performance can only be guaranteed when volumes within the recommended 10...
This Information Applies To: AriaMx Real-time PCR System
Issue
The saturation level is unique for each optical module after the offset of calibration factor and background calibration. The FAM, ROX, HEX, CY3 & CY5 optical modules will saturate at approximately 32,000 RFU while ATTO optical module...