This Information Applies To: Agilent OpenLab CDS Rev. 2.x, Agilent GC Systems
Issue
If a column installed in the GC is used for multiple methods and you want to prevent column configuration changes, you should "lock" the column from the acquisition...
This Information Applies To: Agilent GC Systems
Issue
GC Flame Ionization Detector (FID) Baseline Output stuck at more than 500,000 pA up to 8 million counts.
Background
If the FID interconnect spring is bent, damaged or has bad contact with...
This Information Applies To: All Agilent Gas Chromatographs, 6890, 6850, 7890, 7820, 8890, and 8860.
Issue:
A quick reference guide on the capillary column installation for the GC.
Inlets
Detectors
Split/splitless inlet Purged...
This Information Applies To: Agilent OpenLab CDS (Ver 2.x)
Content:
The following procedure shows how to add a GC instrument in OpenLab CDS (Ver 2.x).
Steps to Follow:
1. Open OpenLab CDS Control Panel, in the Instruments interface, select...
This Information Applies To: Agilent Intuvo 9000, 8860, 8890, 7890A/B, 7820A, 6890, 6850 GC Systems with Split/Splitless, or Multimode Inlets.
Issue:
This article describes how to replace the inlet septum for Split/Splitless and multimode inlets...