Hi all
I'd like to find help with choosing the right liner and syringe for my application.
Following ASTM D5501 rev12 method.
By the way, is there tools for me to find vapor volume or other tricks to find the right liner ?
Thanks
Hi all
I'd like to find help with choosing the right liner and syringe for my application.
Following ASTM D5501 rev12 method.
By the way, is there tools for me to find vapor volume or other tricks to find the right liner ?
Thanks
Hi chesstu,
I have updated the tags on this post. I have also included a link below to the GC calculators.
This GC Calculator package includes a Pressure Flow Calculator, Vapor Volume… |
Here is a link to an application note on the method from Agilent.
https://www.agilent.com/cs/library/applications/5990-6650EN.pdf
As additional information to the method send by Ron Tackett: the column being the 100% PDMS 150m x 0.25mm x 1.0um is for instance the CP-Sil PONA CB in these dimensions and engineered for hydrocarbons…
Here is a link to an application note on the method from Agilent.
https://www.agilent.com/cs/library/applications/5990-6650EN.pdf
Here is a link to an application note on the method from Agilent.
https://www.agilent.com/cs/library/applications/5990-6650EN.pdf