This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG DNA Analyzer, and Oligo Pro II
Issue
Methods for storing capillary array when not in use for longer than two weeks. The array can either be stored installed in the instrument or in the external array docking...
This Information Applies To: Agilent 2100 Bioanalyzer system, single stranded DNA
Issue
Is it possible to run single-stranded DNA (ssDNA) samples on the Bioanalyzer? Are there any Bioanalyzer assays recommended for ssDNA?
Resolution
Agilent does not offer any Bioanalyzer assays dedicated...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Agilent Femto Pulse Systems, Agilent ZAG DNA Analyzer System, Agilent Oligo Pro II, Agilent ProSize data analysis software
Issue
The ProSize software is flickering. It is usually due to incompatible display resolution settings...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG, Oligo Pro II, and ProSize data analysis software
Issue
After the PDF report has been generated in ProSize, the report file is saved with a size of 0 bytes and will not open.
Background
A PDF report can...
This Information Applies To: Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, OligoPro, and ZAG DNA Analyzer
Issue
Perform a capillary alignment whenever a new array has been installed on an Agilent Parallel Capillary Electrophoresis instrument. An alignment is also needed if the capillary...