Replacement for a TV301 NAV with a control unit model 969-8973

Good day, I am currently looking for a replacement for a TV301 NAV without having to change the control unit. The control unit we have installed is a model 969-8973. Can you please send me a catalog of turbo pumps that are compatible with set control unit?

Parents
  • The current model is the 305FS turbo and controller.

    The Agilent turbo and controllers are matched (tuned) and need to be paired together.

     

    The soft start and normal ramp frequencies are tuned to the matching pump to avoid running at resonant frequencies that can cause excessive vibration and bearing wear/damage.

    Using the incorrect turbo controller can damage the pump by ramping and spinning it at resonant speeds that can jeopardize the mechanical integrity (bearing damage).

    In some cases, catastrophically.

Reply
  • The current model is the 305FS turbo and controller.

    The Agilent turbo and controllers are matched (tuned) and need to be paired together.

     

    The soft start and normal ramp frequencies are tuned to the matching pump to avoid running at resonant frequencies that can cause excessive vibration and bearing wear/damage.

    Using the incorrect turbo controller can damage the pump by ramping and spinning it at resonant speeds that can jeopardize the mechanical integrity (bearing damage).

    In some cases, catastrophically.