OpenLAB Custom Calculator

After running a GC sequence which generates data for blanks, standards, method blank and samples - I would like to subtract the area for a particular compound or timed group for the method blank from each of the samples.

I have been using the CC Editor to put custom calculations into the Injection Results tab of the sequence when viewing the data in data analysis (DA).  But I cannot seem to make it subtract one from the other. When you click on a particular injection in the sequence, it only gives you the data for that injection and does not give you the data for the method blank for exmple if you are viewing the sample injection?

Any help would be appreciated.

Parents
  • Hello,

     

    I added tags to this post to increase visibility. 

     

    What revision of OpenLab CDS are you using? 

     

    Is there just going to be one blank run in your sequences or are there going to be multiple blank runs per sequence. If there are going to be multiple blank runs, then do you want to average the areas of your peak of interest across the injections and then subtract that value from the rest of the runs?

     

    And to clarify, you only want to take the area of one compound from the blank run to then subtract from other runs, correct? Or do you want to do this for all compounds? 

  • Hello ryoboyle,

    Thank you for the reply and for realising that I had posted  a duplicate by mistake - apologies.

    To answer your questions:

    • We are running OpenLAB CDS 2.3
    • There is only going to be one Method Blank per sequence
    • Yes, you are correct, I only want take the area for one specific compound/timed group and subtract that from all the sample injections only.

     

    I hope this helps, thanks

Reply
  • Hello ryoboyle,

    Thank you for the reply and for realising that I had posted  a duplicate by mistake - apologies.

    To answer your questions:

    • We are running OpenLAB CDS 2.3
    • There is only going to be one Method Blank per sequence
    • Yes, you are correct, I only want take the area for one specific compound/timed group and subtract that from all the sample injections only.

     

    I hope this helps, thanks