OpenLab CDS Reporting -> extracted chromatograms need to stay same size

Hi,

I'm creating reporting templates but I'm encountering a problem. Whenever there are 1 or 2 peaks with mass in the TIC, the size of the extracted MS peak is fine.

When it is more than 2 it gets too small, how can I adjust the report snippet to stay the size that is in the report template so that whatever amount peaks with MS are integrated it stays the same size?

I'm using openlabs 2.7.

Parents
  • Hello,

    My suggestion would be to use the MS Spectra Flowlayout object. The other option would be to group and repeat the current spectral object by peak retention time.  See the example below, the first is the flowlayout object customized for 2 signals per repeat. The second is the MS Spectrum Overlaid plot grouped with and RT field and repeated on RT. 

    Marty

Reply
  • Hello,

    My suggestion would be to use the MS Spectra Flowlayout object. The other option would be to group and repeat the current spectral object by peak retention time.  See the example below, the first is the flowlayout object customized for 2 signals per repeat. The second is the MS Spectrum Overlaid plot grouped with and RT field and repeated on RT. 

    Marty