Specify ECM/sequence version on report editor for OpenLab Chemstation Edition

The report editor currently has the option for the revision number of report templates at the footer. However, my focus lies on specifying the version of the sequence uploaded onto the ECM this reflects the final version to match with the authorised sequence.

Unfortunately, I haven't yet identified an option for this purpose. If there's a date/time or expression that associates with the saves/uploads for the formatting in the report editor, I would greatly appreciate any advice and suggestions.

Parents
  • Hello,

    What version of Chemstation are you using? You should see in the IR report editor under the live system values the storage fields. In older revisions the word storage was replaced with ECM. Also, in older versions the values would be empty if the report was created in the review or report preview modes. You had to create the report during acquisition or reprocessing for the value to be shown.  

    Marty

Reply
  • Hello,

    What version of Chemstation are you using? You should see in the IR report editor under the live system values the storage fields. In older revisions the word storage was replaced with ECM. Also, in older versions the values would be empty if the report was created in the review or report preview modes. You had to create the report during acquisition or reprocessing for the value to be shown.  

    Marty

Children