I would like to relate the peak retention time of a calibration line to the test retention time. How to Get the Calibration Line Retention Time
THANK YOU FOR YOUR HELP
syb
I would like to relate the peak retention time of a calibration line to the test retention time. How to Get the Calibration Line Retention Time
THANK YOU FOR YOUR HELP
syb
thanks, but this formula will only work if the calibrators are in the resultset.
Can we extract the retention times from the respective calibration curves, in order to be able to use it in Openlab 2.5, thanks for your help
SYB
Hello,
No that is not possible in CC or IR if the calibration standards are not present in the result set. What you could do is setup the method to update the method RT at 100% when running the standards. This way the expected RT is the same as the found calibration standard RTs for each compound. At that point you could just subtract the ERT from the RT to get the delta.
Marty
Hello,
No that is not possible in CC or IR if the calibration standards are not present in the result set. What you could do is setup the method to update the method RT at 100% when running the standards. This way the expected RT is the same as the found calibration standard RTs for each compound. At that point you could just subtract the ERT from the RT to get the delta.
Marty
Thank you, I'll do as you say,
SYB