Batch Records for cGMP facility

Hi, We are a cGMP facility and using Open Lab 2.5 CDS and we are a batch process. We would like to have the CDS system compile all of the chromatography results (about 7 from different instruments) into a single folder with the batch number. Is this possible or something like it?

  • Hello,

    When you say results does that mean raw data or reports? For the raw data you can control the storage location using projects and then subfolders under the project. If you want the reports to go to a central share in a specific folder that is possible using a unc path in the report definition copy to section of the method. You can either hardcode the path in the method or use tokens based on certain criteria. 

    Marty