Using OLCDS 2.4. Is there a way to view, filter, or report datasets that are not electronically signed? We would like to be able to see what datasets have not been signed. The only way to do this currently is by opening each individual dataset.
Using OLCDS 2.4. Is there a way to view, filter, or report datasets that are not electronically signed? We would like to be able to see what datasets have not been signed. The only way to do this currently is by opening each individual dataset.
Hello,
Yes, you can do this in the content manager web interface, but you would need to have CDS 2.5 or higher. You also need to have ECM-XT not ECM 3.x as the backend content manager.
Marty Adams