Can someone explain to me when is preferable to use Assymetric EIC instread of Symmetric EIC?
thank you
Find maintenance, best practice, and troubleshooting articles.
Can someone explain to me when is preferable to use Assymetric EIC instread of Symmetric EIC?
thank you
Hi JF_Fortin
I would use asymmetric when I just want to use an integer mass to extract a chrromatogram and the software allows for a small negative mass defect and a larger positive defect. If I want to use the mass with decimal places, i.e. including the mass defect, then symmetrical extraction makes more sense to me.
/Andy
Thank you Andy. I use EIC for quantitative purpose instead of using TIC. I think based on your response, asymmetric seems a logical choice.
Thank you Andy. I use EIC for quantitative purpose instead of using TIC. I think based on your response, asymmetric seems a logical choice.