Chemstation Reporting Filters

I am sure there is an easy way to accomplish this, and I am probably overthinking it, but how would you filter to only show results from the 260 wavelengths? It is set as signal_ID MWD1A.  However, when I attempt to add the string to the filter the report errors out.  I am sure it is just the formatting of the string filter.  I am attempting to build a simple template to check peak heights of samples, but only of the one wavelength (as we collect at 3 wavelengths). 

Parents
  • Hello,

    What Chemstation version is the for and are you using Intelligent Reporter for the report? If you want to filter on the wavelength number, then use signal description not signal name in the filter. You could also use the signal name but that would be just the MWD1A as the filter expression in that case you can use = instead of contains. 

    Marty

Reply
  • Hello,

    What Chemstation version is the for and are you using Intelligent Reporter for the report? If you want to filter on the wavelength number, then use signal description not signal name in the filter. You could also use the signal name but that would be just the MWD1A as the filter expression in that case you can use = instead of contains. 

    Marty