Report template (Open Lab CDS)

Hello everyone,

Our sequence report template doesn't show the line# in sequential order t(his happens specifically when we run single standard method and choose the same vial location). please see attached. 

/resized-image/__size/1280x960/__key/communityserver-discussions-components-files/26/pastedimage1680557386383v1.png

Could anyone provide their feed back on this please?

Thank you, 

Barinder

  • Hi Barinder,

    The line number is taken from the sequence. When you reinject the same sample (Inj/vial) multiple times, it is still on the same line in the sequence, hence why the order appears to be missing for those replicate injections.

    There should be an Inj# column. Having this next to the Line# may help. If that column is not available, add a column and make sure its Value is: =first(Injection_OrderNo).

  • Hello,

    To get the output you are seeing, hide duplicates must be checked for that column, as I would expect to see two repeated for all injections where it is blank. You can add the injection order as a column as suggested in the previous post. If what you want is a numerical index for the table, you can use a custom calculation as below to create an index and then display it in IR. The same thing can also be accomplished by using a hidden column and storing the Sample_AcquisitionOrderNo as an aggregator, and then displaying the count as the index column. 

    Marty Adams

    CurrentSequence.AllInjections.OrderBy(function(x) x.CreationTime).Select(function(x) x.CreationTime).Distinct.ToList.FindIndex(function(x) x=CurrentInjection.CreationTime)+1

Was this helpful?