Hello,
When I try to create a new nested baseline, the baseline won't go past the middle of the peak. I need to separate an entire peak from a range of the chromatogram, does anybody know how to get around this issue?
Thanks!
Find maintenance, best practice, and troubleshooting articles.
Hello,
When I try to create a new nested baseline, the baseline won't go past the middle of the peak. I need to separate an entire peak from a range of the chromatogram, does anybody know how to get around this issue?
Thanks!
Hi Dylan
Can you share a picture of the peak that you're trying to integrate? Nested baselines can't go beyond the peak apex since they are intended to be front or tail tangent skims, which shouldn't really go beyond the apex of the main peak.
/Andy
Thanks for the response Andy! The peak at 7.822 min. is a surrogate peak and needs to be integrated separately from the range that is in included in.
Hi Dylan
Thanks for sharing the chromatogram, I can see the problem clearly now. I don't think I have a good answer for you, so hopefully, someone else in the community can find an answer for you.
/Andy
Thanks for taking a look
Thanks for taking a look