Problems with automated analysis and license breakdown?

We are running long term tests with multiple GC analysis, automatically controlled from LabVIEW via ActiveX communication with OpenLab EZChrom.

We recently had an event in the EZChrom software creating problems;

suddenly midway through the tests, the analysis started to fail. The communication seemed to fail,

resulting in an error that suggested OpenLab couldn’t locate its core license. The problem was solved by entering off-line EZChrom software to copy license, erase license and ultimately re-paste license. Communication returned, and the test could continue.

 

Anybody seen these strange things that the license system seems to break down during testing ?

Rather hopeless bug to fight with......

 

Martin

Parents
  • Is this a client/server installation or a workstation installation? 

     

    The first thing that I would check after getting an error message would be the license page in the OLCP by going to Administration>Licenses. When the error appears, does the registered license show as being valid or invalid in the OLCP? It would also be worthwhile to verify that both the Agilent OpenLab License Server and Agilent OpenLab Licensing Support services are running on the component of your system that manages licenses (application server for a client/server system and the workstation itself if this is a workstation). 

     

    If the services are running, then the next step would be to click on the 'View' option in the licenses page of the OLCP, which will bring up the licensing web ui. This will show you how many of each license is registered and how many are currently in use. 

  • Its a standalone workstation installation, one computer directly towards one GC.

    I'll notify my customers about your suggestion, might be some time before we get any feedback on this. Thanks so far.

Reply Children
No Data
Was this helpful?