Is there a how-to guide on theoretical plates?
I am stuck like a stick in the mud.
Find maintenance, best practice, and troubleshooting articles.
Is there a how-to guide on theoretical plates?
I am stuck like a stick in the mud.
Hello,
In your method, you should have a System Suitability section where you can setup the calculations. You need to select what peaks you want to calculate system suitability for in the injections. You can also select which calculation types to use EP, JP, or USP. Note: you can make multiple selections. This is all you need to calculate the base theoretical plates results. If you want the plates per meter, then you will need to enter the column information if not set by the instrument acquisition.
Marty Adams
Hello,
In your method, you should have a System Suitability section where you can setup the calculations. You need to select what peaks you want to calculate system suitability for in the injections. You can also select which calculation types to use EP, JP, or USP. Note: you can make multiple selections. This is all you need to calculate the base theoretical plates results. If you want the plates per meter, then you will need to enter the column information if not set by the instrument acquisition.