How to add % recovery to calibration table in OpenLab version C.01.07 SR1

I would like to add some information to the calibration table but am not sure if it is possible.  I would like to add two columns: the calculated concentration of each calibration standard and the % recovery or difference from the true value.  Currently the default table shows the Amount (true value of the standard), the Area in the standard and the Response Factor in addition to other info.  I don't care about the RF and would be happy to remove that if I can get the concentration and % difference added.  Thanks,

Parents Reply
  • Hello,

     

    If you select 'Amount' under the 'Select the value to be calculated' section, then your calibration table will expect you to manually enter a response factor (Amount/Response) which will then be used along with the integration results to calculate the values in the Amount column of the calibration table.

     

    From the ChemStation help, the Response% column calculates the Response % of a peak at this calibration level.