,Hello
I need your help.
Is it possible to have in the same formula a specific calculation for each compound for example:
If peak 1 compound amount *5/sample weight
If peak 2 compound amount *10*/sample weight
Thank you for your help
SYB
,Hello
I need your help.
Is it possible to have in the same formula a specific calculation for each compound for example:
If peak 1 compound amount *5/sample weight
If peak 2 compound amount *10*/sample weight
Thank you for your help
SYB
What about using the compound multipliers in the Calibration part of the processing method with a "mass%" type calibration?
/Andy
What about using the compound multipliers in the Calibration part of the processing method with a "mass%" type calibration?
/Andy
Hello,
You have 3 options for something like this.
1) As Andy said you could use the compound multiplier in the method calibration table. Note: If you use mass% and do not want the result in % you will need to use one of the multiplier or dilution fields to remove that /100
2) Compound custom field in the project.
3) Constants file in the CCF when using custom calculations.
Marty Adams