Report Calibration

Hi Everyone

I am using Agilent 7890B for doing analysis permanent gas via methanizer. I have 3 mix std which have total component 3% and have do calibration for to know average RF.

question is if the RF factor put in manual in openlab chemsation it always show %norm 100%, beside that we only our component totally 3%

Thanks

  • If the report is always normalizing the amounts to total 100%, it sounds like the quantitation settings are not correct.  In the Data Analysis view, go to the Report menu and select Specify Report.  In the dialog box that opens, click on the Quantitation Settings tab.  In the Calculation Mode section, make sure the Calculate drop-down is set to ESTD.